I checked the accuracy of my Fluke 73 multimeter in Volts AC mode using sine waves and found the following; its accurate at lower frequencies (20 to 640 Hz) but gets steadily more inaccurate as frequency increases.
Sine waves were generated using Realtime Analyzer (
http://www.ymec.com , a very useful program!)- fed through mixer to the power amp (bypassing the DRPA). Final voltage for signal was set at 4.00 volts (Peak) and measured via a calibrated Tektronix TAS220 oscilloscope. The scope and the Fluke 73 were connected to the signal simultaneously. Signal voltage on the scope was checked each time the frequency was changed but it showed no noticeable change over the range tested.
Actual Peak Voltage-------- square root of 2-------- Actual RMS voltage
4.00 1.4142 2.828
Tested Points
Frequency (Hz)-------- RMS Volts observed-------- % of Actual
20-------- 2.773-------- 98.0
40-------- 2.828-------- 100.0
80-------- 2.839-------- 100.4
160-------- 2.840-------- 100.4
320-------- 2.838-------- 100.3
640-------- 2.822-------- 99.8
1280-------- 2.756-------- 97.4
2560-------- 2.531-------- 89.5
3000-------- 2.434-------- 86.1
4000-------- 2.199-------- 77.7
5120-------- 1.947-------- 68.8
6000-------- 1.767-------- 62.5
7000-------- 1.583-------- 56.0
8000-------- 1.421-------- 50.2
9000-------- 1.276-------- 45.1
10240-------- 1.122-------- 39.7
(The data fit a 4th order polynomial pretty well - from 320 Hz and above, anyway)
Results:
At 20 Hz the meter dips a bit in accuracy, but 98% of actual isn't too far off. For setting subs using any freq. in the range 40 to 160 would be fine....but it seems prudent to pick a frequency a good bit away from your system's high and low pass settings to avoid any attenuation from the filters. Stan Graves' rule of limiting using frequencies 1/3 octave away from any crossover frequencies seems a great place to start.
For setting limits on tops any freq. in the range 80 to 640 Hz should keep your limit setting accurate, again depending on where your high-pass filter is set. For my meter 1000Hz would show rms voltage about 98.6% -- pretty good for government work.
These are just the results of an older Fluke 73, a non-True RMS meter. YMMV depending on the meter you've got. I imagine a True- RMS meter would fair much better but I don't have one to test.
The upshot of all this is that the error could have a serious impact when setting limiters. For example, if you set an output limit of 30Vac using a 8000 Hz test signal, your system would hit 60Vac when it dished real music with substantial content in the range 125 - 640 Hz, blowing just about any driver we deal with here. An unlikely, near-worst case scenario for sure....but you get the point.
The data would also indicate that its not good to use pink noise (or white noise for that matter) when setting limiters for power amp output. (Don't know if anyone does that - but its been discussed.) Error in the higher frequency ranges could likely cause pink noise to show a much lower RMS reading than actual.
Again, YMMV, but I thought some might find this interesting.
Cheers,
Captain D